What increases signal on apparently clean glass?

Over the years working with self-cleaned coverslips, I’ve noticed that when finding focus on a PBS blank, I can obtain a baseline signal value as low as 0.05 and up to over 0.10 on the same slide, even though the FOV in all wells appears to be over clean glass in the native view. This occurs in the absence of any obvious artifact, such as diagonal lines as from well leakage. In the corresponding ratiometric view, the FOV simply appears noisier at the higher signal values. As I am using the same PBS for all wells, I have to believe glass itself is causing the elevated signal, even in the absence of an obvious contaminant. Can you explain how the glass itself, in the absence of unwanted particles in solution or on the glass surface, can elevate background signal? What about the glass might be doing this?

107 Views

Dear Chase,


I’m Benjamin Cappe, Product Manager at Refeyn.

Thanks for the question.


What you're seeing is most likely due to subtle variations at the glass-liquid interface rather than visible contamination. Mass photometry is highly sensitive, and even when a field of view appears clean in the native image, microscopic differences in glass surface properties, batch-to-batch variation, residual contamination, surface roughness, or other locally adsorbed material can increase the background signal and make the ratiometric image appear noisier.


When using self-cleaned coverslips from third parties, it can be difficult to pinpoint the exact source, as factors such as glass batch variability, cleaning procedures, handling, buffers, and plasticware can all contribute to background noise.


This is one reason we developed MassGlass™ UC. These slides are specifically selected by our team, cleaned, and quality-controlled to provide a more consistent glass surface for mass photometry measurements, thereby reducing variability between experiments.


If you'd like to discuss your specific observations in more detail, feel free to contact our support team at support@refeyn.com. We'd be happy to review the data and help troubleshoot.


Best regards,

Benjamin Cappe, Product Manager

Mass calibration with poly-L-lysine slides

Hi, I'm having trouble creating a mass calibration when using slides coated in poly-L-lysine (for r-loop + protein interaction work). BSA, beta amylase, and thyroglobulin are sticking to the slides and I've had some success with beta gal. Are there any protein recommendations?

205 Views
Amy Chau
26 feb

Hi Nicole,


Can I ask what you mean by sticking to the slides? We measure the particle signals as they bind to the slide. If you're not see individual particles during the recording, the concentration might be off. Try using 5-10 nM of protein for measurement


It might be easier to look at the data and provide better troubleshooting help. Feel free to email us at support@refeyn.com or to me directly at amy.chau@refeyn.com. We can discuss more in detail via email or call.


Amy Chau

Sr. Field Applications Scientist

Laser wavelength used in the (macro) mass photometers?

What is the laser wavelength used in the mass photometry instruments including KaritroMP? Is the wavelength the same in all your instruments?


Thank you,

Jitka

133 Views

Directions for using/storing MassFerence P1 Calibrant

Is there any documentation for this?


I read on the vial that the calibrant is 500X and a volume of 50 uL. Can I prepare 20 uL aliquots at 2X and store them at -20 to avoid repeated freeze/thaw cycles?


Any information is welcome.


Thanks,

Ryan

124 Views
Team Refeyn
Team Refeyn
03 mar 2025

Hi Ryan,


We have not tested “20 uL aliquots at 2X” and hence unable to comment on the product stability at such dilutions. But what we can say is for best results, aliquot 10 x 5 µL upon first thaw and store to avoid repeated freeze/thaw cycles.


Best,

Team Refeyn